SCYM Instrument Operation 2 β Questions and Answers
Question 1: What is the primary purpose of the sheath fluid in a flow cytometer?
- To lyse red blood cells before analysis
- To hydrodynamically focus cells into a single-file stream (Correct answer)
- To fluorescently label surface antigens
- To cool the laser during operation
Correct answer: To hydrodynamically focus cells into a single-file stream
Sheath fluid creates laminar flow that hydrodynamically focuses the sample stream so cells pass through the laser interrogation point one at a time.
Question 2: Which parameter is used to distinguish doublets from single cells during flow cytometric analysis?
- FL1 area vs FL2 area
- FSC height vs FSC area (or width) (Correct answer)
- SSC height vs FSC height
- FL3 area vs SSC area
Correct answer: FSC height vs FSC area (or width)
Doublet discrimination is performed by comparing FSC-H (height) vs FSC-A (area); doublets produce disproportionately higher area relative to height.
Question 3: When performing daily QC on a flow cytometer using fluorescent beads, a sudden rightward shift in bead peak CV is most likely caused by:
- Laser power fluctuation (Correct answer)
- Incorrect scatter voltage settings
- Clogged sample line reducing flow rate
- Expired sheath fluid
Correct answer: Laser power fluctuation
Laser power instability causes variable fluorescence excitation, broadening the bead peak and increasing CV.
Question 4: In a spectral flow cytometer, unmixing algorithms require which reference material?
- Unstained cells only
- Single-color controls (single-stained references) for each fluorochrome (Correct answer)
- A compensation matrix from a previous run
- FMO controls for every fluorochrome
Correct answer: Single-color controls (single-stained references) for each fluorochrome
Spectral unmixing requires single-color reference spectra for each fluorochrome to mathematically separate overlapping emission signals across detectors.
Question 5: The time delay setting in a cell sorter is used to:
- Set the interval between successive droplet formations
- Synchronize the deflection charge with the droplet containing the target cell (Correct answer)
- Control the sample injection pressure
- Adjust the laser pulse frequency
Correct answer: Synchronize the deflection charge with the droplet containing the target cell
The time delay (sort delay) accounts for the transit time from the laser interrogation point to the droplet break-off point so the correct droplet is charged.
Question 6: Which action should be taken first if the sample flow rate on a benchtop cytometer drops unexpectedly mid-run?
- Replace the sample tube immediately
- Run the instrument cleaning protocol
- Check for a clog in the sample line and perform a backflush if needed (Correct answer)
- Increase sheath pressure by 10 PSI
Correct answer: Check for a clog in the sample line and perform a backflush if needed
A sudden flow rate drop typically indicates a partial clog; checking and performing a backflush resolves most sample-line obstructions before escalating.
Question 7: What does a low coefficient of variation (CV) for a fluorescent bead peak indicate during instrument QC?
- The detector PMT voltage is set too high
- The instrument is performing consistently with tight measurement precision (Correct answer)
- Excessive background fluorescence is present
- The laser is operating below target power
Correct answer: The instrument is performing consistently with tight measurement precision
A low CV on bead peaks indicates that the instrument is producing highly reproducible fluorescence measurements, reflecting good optical and electronic performance.
What is the primary purpose of the sheath fluid in a flow cytometer?