Computed Tomography Test Computed Tomography Ultimate 4 — Questions and Answers
Question 1: What is the purpose of 'prospective triggering' in cardiac CT?
- To scan continuously throughout the cardiac cycle
- To acquire data only during a specific, predetermined phase of the cardiac cycle (Correct answer)
- To automatically adjust kVp based on heart rate
- To reconstruct multiple cardiac phases from one acquisition
Correct answer: To acquire data only during a specific, predetermined phase of the cardiac cycle
Prospective triggering uses the ECG signal to trigger X-ray exposure only during a specific quiescent phase, significantly reducing radiation dose compared to retrospective gating.
Question 2: Which ionizing radiation unit is used to express the effective dose that accounts for tissue radiosensitivity?
- Gray (Gy)
- Sievert (Sv) (Correct answer)
- Roentgen (R)
- Rad
Correct answer: Sievert (Sv)
The Sievert (Sv) is used to express effective dose, which weights absorbed dose by tissue-specific radiation sensitivity and radiation type weighting factors.
Question 3: In CT angiography, what is the bolus tracking technique used for?
- Reducing the volume of contrast needed
- Triggering the diagnostic scan when contrast reaches the target vessel (Correct answer)
- Measuring blood flow velocity
- Calculating vessel diameter automatically
Correct answer: Triggering the diagnostic scan when contrast reaches the target vessel
Bolus tracking monitors contrast enhancement in a target vessel using repeated low-dose images and automatically triggers the diagnostic scan when the preset HU threshold is reached.
Question 4: What is the main disadvantage of increasing pitch in helical CT?
- Increased radiation dose
- Reduced z-axis resolution (Correct answer)
- Longer scan time
- Decreased detector coverage
Correct answer: Reduced z-axis resolution
Higher pitch values interpolate data across a wider z-axis distance, which degrades longitudinal (z-axis) spatial resolution and can increase image noise.
Question 5: What type of CT artifact is caused by patient movement during data acquisition?
- Ring artifact
- Star artifact
- Motion artifact (Correct answer)
- Cupping artifact
Correct answer: Motion artifact
Motion artifacts appear as blurring, streaks, or ghosting on images when the patient moves during the scan, causing inconsistent projection data.
Question 6: Which window setting is most appropriate for evaluating pulmonary nodules on chest CT?
- Soft tissue window (W:400, L:40)
- Lung window (W:1500, L:-600) (Correct answer)
- Bone window (W:2000, L:500)
- Brain window (W:80, L:35)
Correct answer: Lung window (W:1500, L:-600)
Lung window settings use a wide window width and low center level to display the full dynamic range of pulmonary parenchyma including subtle nodules and airways.
Question 7: What is the primary determinant of contrast-to-noise ratio (CNR) in CT imaging?
- Reconstruction kernel
- Radiation dose (Correct answer)
- Field of view
- Gantry rotation speed
Correct answer: Radiation dose
Radiation dose is the primary determinant of CNR because higher dose reduces quantum noise, making differences in tissue attenuation more detectable.
What is the purpose of 'prospective triggering' in cardiac CT?