CMT Laboratory Quality Control and Management 1 — Questions and Answers
Question 1: What is the primary purpose of running quality control (QC) samples in a clinical laboratory?
- To calibrate analytical instruments
- To monitor the precision and accuracy of test results (Correct answer)
- To train new laboratory staff on procedures
- To document and archive patient results
Correct answer: To monitor the precision and accuracy of test results
QC samples are run to monitor the precision and accuracy of analytical processes and detect errors before patient results are reported.
Question 2: Which Westgard rule is triggered when a single control value exceeds the mean ± 3 standard deviations, indicating a random error?
- 2₂s rule
- R₄s rule
- 1₃s rule (Correct answer)
- 4₁s rule
Correct answer: 1₃s rule
The 1₃s rule is a rejection rule triggered when one control observation exceeds the mean ± 3 SD, indicating a possible random error.
Question 3: In laboratory quality management, what does 'accuracy' specifically refer to?
- The reproducibility of repeated measurements under the same conditions
- How close a measurement is to the true or accepted reference value (Correct answer)
- The range of values that includes 95% of a reference population
- The lowest analyte concentration an instrument can reliably detect
Correct answer: How close a measurement is to the true or accepted reference value
Accuracy refers to how closely a measured value agrees with the true or accepted reference value.
Question 4: What does the coefficient of variation (CV) primarily assess in laboratory quality control?
- Accuracy of a method compared to a reference method
- Bias between two analytical methods
- Imprecision (random error) of an analytical method (Correct answer)
- The linearity of an assay across its reportable range
Correct answer: Imprecision (random error) of an analytical method
The coefficient of variation (CV = SD/mean × 100%) is a measure of imprecision or random error expressed as a percentage of the mean.
Question 5: Which federal agency administers and enforces CLIA regulations for clinical laboratory quality standards in the United States?
- The Joint Commission (TJC)
- College of American Pathologists (CAP)
- Centers for Medicare & Medicaid Services (CMS) (Correct answer)
- American Society for Clinical Pathology (ASCP)
Correct answer: Centers for Medicare & Medicaid Services (CMS)
The Centers for Medicare & Medicaid Services (CMS) administers and enforces CLIA regulations that establish quality standards for clinical laboratories.
Question 6: When a Westgard 2₂s rule is triggered (two consecutive controls exceed ±2 SD in the same direction), what type of error is most likely present?
- Random error
- Systematic error (Correct answer)
- Gross error
- Pre-analytical error
Correct answer: Systematic error
The 2₂s rule, where two consecutive controls exceed ±2 SD in the same direction, is indicative of systematic error such as a shift or drift in the method.
Question 7: What is the primary purpose of proficiency testing (PT) in clinical laboratories?
- To evaluate individual laboratory employee competency
- To compare a laboratory's performance with peer laboratories using unknown samples (Correct answer)
- To validate new equipment before it is placed into service
- To establish reference ranges for newly implemented assays
Correct answer: To compare a laboratory's performance with peer laboratories using unknown samples
Proficiency testing involves external agencies sending unknown samples to laboratories so their performance can be compared against peer laboratories and accepted values.