CBET General Practice — Questions and Answers
Question 1: A biomedical technician is troubleshooting an infusion pump that delivers inconsistent flow rates. Which component is MOST likely responsible?
- Occlusion sensor
- Peristaltic pump mechanism (Correct answer)
- AC power supply
- LCD display module
Correct answer: Peristaltic pump mechanism
The peristaltic pump mechanism directly controls fluid delivery, and wear or misalignment causes inconsistent flow rates.
Question 2: According to NFPA 99, the maximum allowable ground resistance for a receptacle in a patient care area is:
- 0.5 ohms
- 0.1 ohms (Correct answer)
- 1.0 ohms
- 0.2 ohms
Correct answer: 0.1 ohms
NFPA 99 requires receptacle ground resistance to not exceed 0.1 ohms in patient care areas to ensure electrical safety.
Question 3: Which type of fire extinguisher is appropriate for electrical equipment fires in a biomedical equipment area?
- Class A (water)
- Class B (foam)
- Class C (CO2 or dry chemical) (Correct answer)
- Class D (dry powder)
Correct answer: Class C (CO2 or dry chemical)
Class C extinguishers use non-conductive agents like CO2 or dry chemical, making them safe for energized electrical equipment fires.
Question 4: A patient monitor displays an artifact that disappears when the patient's lead wires are disconnected. This MOST likely indicates:
- A faulty amplifier in the monitor
- EMI from external sources coupling through the leads (Correct answer)
- A defective power supply
- Software corruption in the monitor
Correct answer: EMI from external sources coupling through the leads
When artifact disappears upon lead disconnection, the leads are acting as antennas picking up electromagnetic interference from external sources.
Question 5: The term 'microshock' refers to hazardous current delivered:
- Through intact skin at levels above 100 mA
- Directly to the heart via an intracardiac catheter at levels as low as 10 µA (Correct answer)
- Through a defibrillator paddle during cardioversion
- Via a transcutaneous electrical nerve stimulator
Correct answer: Directly to the heart via an intracardiac catheter at levels as low as 10 µA
Microshock occurs when current bypasses skin resistance and reaches the heart directly, where as little as 10 µA can induce ventricular fibrillation.
Question 6: During preventive maintenance on a ventilator, the BIOMED finds the exhalation valve is slow to close. The MOST immediate patient safety concern is:
- Increased FiO2 delivery
- Inadequate tidal volume delivery due to gas leakage (Correct answer)
- Elevated baseline airway pressure
- Reduced alarm sensitivity
Correct answer: Inadequate tidal volume delivery due to gas leakage
A slow-closing exhalation valve allows delivered gas to escape before full tidal volume is delivered to the patient's lungs.
Question 7: Which standard specifically governs the safety requirements for medical electrical equipment in the United States?
- IEC 60601-1
- ANSI/AAMI ES60601-1 (Correct answer)
- UL 2601-1
- NFPA 70E
Correct answer: ANSI/AAMI ES60601-1
ANSI/AAMI ES60601-1 is the US adoption of IEC 60601-1 and governs general safety requirements for medical electrical equipment.
Question 8: A hospital's isolated power system (IPS) is designed primarily to protect against:
- Power surges from the utility grid
- Macroshock from single-fault ground currents in wet locations (Correct answer)
- Electromagnetic interference from adjacent equipment
- Overloading of the facility's main circuit breakers
Correct answer: Macroshock from single-fault ground currents in wet locations
An IPS isolates the patient care circuit from ground so a single fault does not create a complete circuit for dangerous ground currents (macroshock).
Question 9: When testing a line isolation monitor (LIM), the alarm should activate when the hazard current exceeds:
- 1 mA
- 2 mA
- 5 mA (Correct answer)
- 10 mA
Correct answer: 5 mA
NFPA 99 requires the LIM alarm to activate when total hazard current reaches 5 mA, indicating the IPS isolation has degraded to a dangerous level.
Question 10: A technician notices the hospital's oxygen wall outlet delivers flow but at a lower pressure than specification. The MOST likely cause is:
- Contamination of the oxygen supply with nitrogen
- A partially closed or faulty zone valve upstream
- A calibration error in the flowmeter
- Excessive demand from too many simultaneous users on that zone (Correct answer)
Correct answer: Excessive demand from too many simultaneous users on that zone
Excessive simultaneous demand on a piped medical gas zone creates a pressure drop due to flow resistance in the supply piping.
Question 11: In a digital-to-analog converter (DAC) used in medical equipment, increasing the bit resolution from 8-bit to 12-bit:
- Doubles the output voltage range
- Increases the number of discrete output levels from 256 to 4,096 (Correct answer)
- Reduces conversion speed by a factor of 4
- Decreases the signal-to-noise ratio
Correct answer: Increases the number of discrete output levels from 256 to 4,096
An 8-bit DAC produces 2^8 = 256 levels while a 12-bit DAC produces 2^12 = 4,096 levels, providing finer output resolution.
Question 12: A technician is performing annual PM on a hospital bed. Which safety check is MOST critical for patient fall prevention?
- Verifying the bed's Trendelenburg angle accuracy
- Testing siderail locking mechanisms for secure engagement (Correct answer)
- Calibrating the bed's weight scale
- Inspecting the mattress cover for tears
Correct answer: Testing siderail locking mechanisms for secure engagement
Siderail locking mechanism failure is directly linked to patient falls, making it the highest-priority safety check during hospital bed PM.
Question 13: The International System of Units (SI) base unit for electric current is the:
- Volt
- Watt
- Ampere (Correct answer)
- Coulomb
Correct answer: Ampere
The ampere (A) is the SI base unit for electric current, defined as the flow of electric charge past a point per unit time.
Question 14: A biomedical technician must ensure an ECG machine's patient lead leakage current does not exceed which limit for a CF-type device per IEC 60601-1?
- 100 µA under normal conditions
- 10 µA under normal conditions (Correct answer)
- 50 µA under normal conditions
- 500 µA under normal conditions
Correct answer: 10 µA under normal conditions
For CF (cardiac floating) applied parts, IEC 60601-1 limits patient leakage current to 10 µA under normal conditions due to direct cardiac risk.
Question 15: When using a multimeter to measure AC voltage on a 120V circuit, the technician should set the range to:
- The lowest range below 120V for best accuracy
- A range equal to or above 120V to prevent meter damage (Correct answer)
- DC voltage at 200V
- Resistance mode to check the circuit first
Correct answer: A range equal to or above 120V to prevent meter damage
Always select a range at or above the expected voltage to prevent damage to the meter and ensure safe, accurate measurement.
Question 16: A hospital laboratory centrifuge vibrates excessively during operation. The MOST likely cause is:
- Motor bearing wear causing rotor imbalance
- An unbalanced load due to improperly paired or misplaced samples (Correct answer)
- Excessive ambient temperature in the lab
- A failed tachometer sensor
Correct answer: An unbalanced load due to improperly paired or misplaced samples
Centrifuges must have balanced loads; unequal or improperly positioned samples create centrifugal force imbalance causing severe vibration.
Question 17: Which type of radiation is MOST effectively stopped by a thin sheet of aluminum foil?
- Gamma rays
- X-rays
- Beta particles (Correct answer)
- Alpha particles
Correct answer: Beta particles
Beta particles (electrons) have moderate penetration and are effectively shielded by a few millimeters of aluminum, unlike alpha (paper stops it) or gamma (requires dense shielding).
Question 18: A technician completing a work order should document which of the following to satisfy Joint Commission traceability requirements?
- Only the parts replaced and their cost
- Device identification, service performed, test results, technician ID, and date (Correct answer)
- The patient census count at time of service
- Only failures found, not routine PM results
Correct answer: Device identification, service performed, test results, technician ID, and date
Joint Commission requires complete service records including device ID, work performed, test results, technician identification, and the service date for full traceability.
Question 19: The output impedance of a device should ideally be low because:
- It increases the device's power consumption
- It minimizes signal voltage drop when driving a load (Correct answer)
- It prevents electromagnetic interference
- It increases current draw from the power supply
Correct answer: It minimizes signal voltage drop when driving a load
Low output impedance ensures minimal voltage drop across the source impedance when current flows, maintaining signal integrity to the load.
Question 20: Which sterilization method is MOST appropriate for heat-sensitive flexible endoscopes?
- Steam autoclave at 134°C
- Ethylene oxide (EtO) gas sterilization (Correct answer)
- Dry heat oven at 160°C
- UV light irradiation
Correct answer: Ethylene oxide (EtO) gas sterilization
Ethylene oxide gas sterilizes at low temperatures (37–63°C), making it suitable for heat-sensitive devices like flexible endoscopes that would be damaged by steam or dry heat.
Question 21: In an op-amp configured as an inverting amplifier with Rf = 100 kΩ and Rin = 10 kΩ, the voltage gain is:
- +10
- -10 (Correct answer)
- +0.1
- -0.1
Correct answer: -10
The inverting amplifier gain is -Rf/Rin = -100kΩ/10kΩ = -10, where the negative sign indicates phase inversion.
A biomedical technician is troubleshooting an infusion pump that delivers inconsistent flow rates.
Which component is MOST likely responsible?
What Practice Test Geeks adds beyond an AI chat for CBET
An AI chatbot can quiz you, but it can't reproduce the exam conditions or keep your record. Here you get:
- A timed, full-length CBET mock exam — 165 questions in 210 minutes with a 68% pass mark, modeled on the official exam's question count and time limit, followed by a score report broken down by exam domain.
- A fixed CBET question bank: the same test every attempt, so your score is comparable from one try to the next instead of changing with every regenerated quiz.
- Progress saved on your device: completed tests, scores and missed questions stay in My Progress, so you can retake only what you got wrong.
- A printable CBET cheat sheet you can print or save as a PDF.
- CBET flashcards built from the same question bank.
These practice questions were written with AI tools. They are not real exam questions. Exam facts are checked against the official sources listed below. Not reviewed by a subject-matter expert. Editorial policy