RMGA Installation Codes & Standards Compliance 2 — Questions and Answers
Question 1: According to NFPA 99, what is the minimum required pressure for a Level 1 medical oxygen system at the outlet?
- 345 kPa (50 psi)
- 414 kPa (60 psi) (Correct answer)
- 275 kPa (40 psi)
- 483 kPa (70 psi)
Correct answer: 414 kPa (60 psi)
NFPA 99 requires Level 1 medical oxygen systems to maintain a minimum of 414 kPa (60 psi) at the station outlet.
Question 2: Under CGA G-4.1, which material is prohibited for use in oxygen service piping due to flammability risk?
- Type K copper tubing
- Stainless steel 316L
- Polyvinyl chloride (PVC) (Correct answer)
- Brass fittings
Correct answer: Polyvinyl chloride (PVC)
CGA G-4.1 prohibits PVC and other combustible plastics in oxygen piping systems because they can ignite under high-pressure oxygen environments.
Question 3: What does NFPA 99 classify as a 'Category 1' gas system risk level?
- Systems where failure is unlikely to cause injury
- Systems where failure poses a risk of major injury or death (Correct answer)
- Systems used only for diagnostic purposes
- Systems with redundant backup supply
Correct answer: Systems where failure poses a risk of major injury or death
NFPA 99 Category 1 systems are those where equipment failure could result in serious injury or death to patients or caregivers.
Question 4: Per ASSE 6010 standards, who is qualified to perform work on medical gas and vacuum piping systems?
- Any licensed plumber with 2 years of experience
- A medical gas installer certified to ASSE 6010 (Correct answer)
- A mechanical engineer registered in the state
- Any HVAC technician with medical facility experience
Correct answer: A medical gas installer certified to ASSE 6010
ASSE 6010 certification is specifically required for personnel who install, inspect, and maintain medical gas and vacuum piping systems.
Question 5: According to NFPA 99, what is the maximum allowable pressure drop across a medical gas zone valve during peak flow conditions?
- 35 kPa (5 psi)
- 70 kPa (10 psi)
- 14 kPa (2 psi) (Correct answer)
- 105 kPa (15 psi)
Correct answer: 14 kPa (2 psi)
NFPA 99 limits pressure drop across zone valves to 14 kPa (2 psi) at peak demand to ensure adequate pressure at point of use.
Question 6: Which standard governs the color coding of medical gas piping in the United States?
- ANSI Z535.1
- CGA C-9 (Correct answer)
- ISO 32
- ASME B31.1
Correct answer: CGA C-9
CGA C-9 establishes the standard color coding for medical gas piping, with specific colors assigned to each gas type.
Question 7: What action is required when a new medical gas system segment is added to an existing NFPA 99 compliant system?
- Notify the AHJ within 30 days of completion
- Perform verification testing on the new segment only (Correct answer)
- Perform verification testing on the entire system
- Submit as-built drawings to the hospital engineering department
Correct answer: Perform verification testing on the new segment only
NFPA 99 requires verification testing of only the newly added segment, not the entire existing system, when making additions.
According to NFPA 99, what is the minimum required pressure for a Level 1 medical oxygen system at the outlet?