EDAIC Anesthetic Equipment and Safety 4 — Questions and Answers
Question 1: Which component of the anesthesia machine is responsible for reducing cylinder pressure from approximately 2000 psi to a safe working pressure of around 45 psi?
- Pressure regulator (reducing valve) (Correct answer)
- Check valve
- Flow control valve
- Pressure relief valve
Correct answer: Pressure regulator (reducing valve)
The pressure regulator (reducing valve) steps down high cylinder pressure to a safe, consistent working pressure of approximately 45 psi for use in the machine.
Question 2: A patient breathing circuit shows persistently elevated CO2 on capnography despite adequate fresh gas flow. The MOST likely cause is:
- Exhausted CO2 absorbent (Correct answer)
- Oxygen analyzer malfunction
- Disconnected oxygen flush valve
- Faulty inspiratory unidirectional valve
Correct answer: Exhausted CO2 absorbent
Exhausted CO2 absorbent loses its capacity to neutralize exhaled CO2, leading to rebreathing and elevated inspired CO2 on capnography.
Question 3: When using a Bain circuit (Mapleson D modification), the recommended fresh gas flow for spontaneously breathing adult patients is approximately:
- 2–3 times minute ventilation (Correct answer)
- 1 L/min above tidal volume
- Equal to tidal volume
- 0.5 times minute ventilation
Correct answer: 2–3 times minute ventilation
For spontaneous ventilation with a Bain circuit, fresh gas flows of 2–3 times the minute ventilation (~150–200 mL/kg/min) are required to prevent rebreathing.
Question 4: The Pin Index Safety System (PISS) on gas cylinders is designed to:
- Prevent accidental connection of wrong gas cylinders to yokes (Correct answer)
- Regulate pressure from the cylinder
- Indicate the remaining volume of gas
- Prevent backflow of gases into the cylinder
Correct answer: Prevent accidental connection of wrong gas cylinders to yokes
PISS uses unique pin configurations for each gas to prevent incorrect cylinders from being mounted on the wrong yoke connection.
Question 5: During anesthesia machine checkout, the 'oxygen flush valve' is tested. What pressure does activating this valve deliver oxygen at?
- 35–75 psi (Correct answer)
- 15–30 psi
- 5–15 psi
- 80–120 psi
Correct answer: 35–75 psi
The oxygen flush valve delivers oxygen directly from pipeline supply at 35–75 psi, bypassing the flowmeters and vaporizer.
Question 6: A rotameter (Thorpe tube) in an anesthesia machine is calibrated for a specific gas because flow readings depend on:
- Gas density and viscosity (Correct answer)
- Gas temperature and pressure only
- Cylinder pressure of the gas
- Color-coding of the flowmeter tube
Correct answer: Gas density and viscosity
Rotameter readings depend on the physical properties (density and viscosity) of the gas, so each flowmeter tube is gas-specific and cannot be interchanged.
Question 7: In a circle breathing system, the purpose of the unidirectional (one-way) valves is to:
- Ensure gas flows in one direction, separating inspiratory and expiratory limbs (Correct answer)
- Prevent oxygen from entering the expiratory limb
- Control the rate of CO2 absorption
- Regulate the APL valve pressure
Correct answer: Ensure gas flows in one direction, separating inspiratory and expiratory limbs
Unidirectional valves in a circle system ensure that gas flows only in the correct direction, keeping fresh gas and exhaled gas separated to prevent rebreathing.
Which component of the anesthesia machine is responsible for reducing cylinder pressure from approximately 2000 psi to a safe working pressure of around 45 psi?