CABT Biomedical Equipment Technology & Function 5 — Questions and Answers
Question 1: When testing ground wire integrity on patient care equipment, the AAMI ES1 standard specifies a maximum resistance of:
- 0.5 ohm
- 0.1 ohm (Correct answer)
- 1.0 ohm
- 0.01 ohm
Correct answer: 0.1 ohm
AAMI ES1 requires that the resistance from the grounding pin to any exposed conductive patient-contact surface not exceed 0.1 ohm to ensure effective fault current dissipation.
Question 2: A neonatal incubator's airflow system is designed to maintain what primary environmental parameter for the infant?
- Oxygen concentration above 30%
- Neutral thermal environment by controlling air temperature (Correct answer)
- Humidity above 80% at all times
- Carbon dioxide below 1000 ppm
Correct answer: Neutral thermal environment by controlling air temperature
The incubator's primary function is to maintain a neutral thermal environment (NTE) that minimizes the infant's metabolic oxygen consumption and heat loss.
Question 3: In a hemodialysis machine, the conductivity monitor of the dialysate circuit is critical because it detects:
- Blood leak into the dialysate
- Incorrect electrolyte concentration that could cause patient harm (Correct answer)
- Air bubbles in the blood circuit
- Transmembrane pressure exceeding safe limits
Correct answer: Incorrect electrolyte concentration that could cause patient harm
Conductivity correlates directly with dissolved electrolyte (sodium) concentration; a conductivity alarm stops dialysate delivery to prevent hypo- or hypernatremia.
Question 4: The purpose of a thermal compensation resistor (or thermistor) in a pressure transducer is to:
- Amplify low-amplitude pressure signals
- Correct for sensitivity drift caused by temperature changes (Correct answer)
- Provide electrical isolation from the patient
- Balance the Wheatstone bridge at initial calibration only
Correct answer: Correct for sensitivity drift caused by temperature changes
Strain-gauge pressure transducers are temperature-sensitive; a thermistor or compensation resistor adjusts the bridge output to maintain accuracy across a range of operating temperatures.
Question 5: Which FDA device classification requires the most rigorous pre-market approval (PMA) process, typically applied to life-sustaining or life-supporting devices?
- Class I
- Class III (Correct answer)
- Class II
- Class IIb
Correct answer: Class III
FDA Class III devices (e.g., implantable pacemakers, ventricular assist devices) pose the greatest risk and require PMA with clinical evidence of safety and effectiveness.
Question 6: A hospital's medical gas alarm panel shows a 'low pressure' warning for the piped oxygen system. The FIRST action a BMET should take is:
- Switch patients to portable oxygen cylinders immediately
- Verify the alarm by checking the zone valve pressure gauge and contacting the facility engineer (Correct answer)
- Shut down all equipment connected to the oxygen wall outlets
- Increase the regulator setpoint on the bulk oxygen supply
Correct answer: Verify the alarm by checking the zone valve pressure gauge and contacting the facility engineer
Confirming the alarm with a gauge reading and notifying facilities engineering ensures the response is appropriate and that clinical staff take corrective action if a true emergency exists.
Question 7: Electromagnetic interference (EMI) from a nearby electrosurgical unit most commonly affects which monitoring device by inducing artifact that mimics the signal of interest?
- Pulse oximeter plethysmographic waveform
- ECG rhythm display (Correct answer)
- End-tidal CO2 waveform
- Non-invasive blood pressure reading
Correct answer: ECG rhythm display
ESU RF energy couples into ECG lead wires, which act as antennas, producing artifact that obscures the cardiac signal during electrosurgical activation.
When testing ground wire integrity on patient care equipment, the AAMI ES1 standard specifies a maximum resistance of: