ARDMS SPI Color Doppler and Artifacts 2 — Questions and Answers
Question 1: Which parameter, when increased, most effectively reduces color Doppler aliasing without changing the PRF?
- Increase color gain
- Shift the baseline downward (Correct answer)
- Decrease wall filter
- Increase ensemble length
Correct answer: Shift the baseline downward
Shifting the baseline extends the Nyquist limit in one direction, allowing higher velocities to be displayed without aliasing.
Question 2: A color Doppler image shows a mosaic pattern within a vessel lumen. This most likely indicates:
- Laminar flow at high velocity
- Turbulent or disturbed flow with aliasing (Correct answer)
- Color bleed from adjacent tissue
- Incorrect wall filter setting
Correct answer: Turbulent or disturbed flow with aliasing
A mosaic pattern of mixed colors represents turbulent flow combined with aliasing, commonly seen distal to a stenosis.
Question 3: Which of the following best describes the color Doppler 'twinkling artifact'?
- Rapidly alternating red and blue pixels behind a strongly reflective surface (Correct answer)
- Color signal appearing in an avascular structure
- Bidirectional flow displayed at a valve
- Acoustic shadowing with color dropout
Correct answer: Rapidly alternating red and blue pixels behind a strongly reflective surface
The twinkling artifact produces rapidly flickering red and blue pixels posterior to calcifications or stones due to phase jitter.
Question 4: When the color Doppler gate is positioned at 90° to flow direction, the resulting color display will show:
- Maximum color saturation
- Aliasing
- No color signal (Correct answer)
- Reversed color assignment
Correct answer: No color signal
At 90° insonation angle the Doppler frequency shift is zero, so no color signal is generated.
Question 5: Color Doppler flash artifact is characterized by:
- Persistent color signal in a cyst
- Transient color fill of entire image from tissue motion (Correct answer)
- Color dropout in a high-velocity jet
- Reverberant color stripes
Correct answer: Transient color fill of entire image from tissue motion
Flash artifact occurs when patient or probe motion generates low-frequency Doppler shifts that briefly fill the color box with color.
Question 6: Which color Doppler parameter directly controls the number of pulses used to estimate each color pixel?
- Color gain
- Wall filter
- Ensemble length (packet size) (Correct answer)
- Color priority
Correct answer: Ensemble length (packet size)
Ensemble length (packet size) is the number of pulses per scan line used for color velocity estimation; more pulses improve sensitivity but reduce frame rate.
Question 7: A color Doppler image of the carotid artery shows color signal extending beyond the vessel wall into surrounding soft tissue. This is called:
- Color bleed (blooming) (Correct answer)
- Twinkling artifact
- Mirror image artifact
- Color aliasing
Correct answer: Color bleed (blooming)
Color bleed or blooming occurs when gain is set too high, causing color to spill beyond true vessel margins.
Which parameter, when increased, most effectively reduces color Doppler aliasing without changing the PRF?